KH reference methods, the drop checker and the pH probe adaptations

Discussion in 'CO2 Enrichment' started by Tom Barr, Dec 4, 2006.

  1. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    This is the thread to place the various adaptations of Vaughn's reference idea for the drop checker and the pH reference probe based upon the same principle.

    The goal of using a KH reference solution is both simple and elegant.
    Many folks have variations and issues determining their KH in terms of precise amount and one without interferences from other things such as tannins, peat, ADA aqua soil among other issues such as anythuing that influences non carbonate alkalinity and other acids than CO2 gases.

    There are a few issues bu they are resolved and almost done with the pH probe.

    The purpose here is to detail out each method and the various trade offs and advantages each has. They are essentaially the same as the pH colormetric cheapy test and that of the pH probe for measuring the pH against a a Kh test measurement.

    The issue has not been so much the pH test, rather, the KH.

    Vaughn asks from the APD:

    "So, do you envision this as a product we would buy from the
    probe
    manufacturer - A CO2 probe?"

    Sort of.


    " The only downside I see to that is that
    it offends my DIY sensibilities, but I do notice that I haven't
    been
    trying to make my own pH probe either, so I can hardly object."

    It's no different than what we have today, the DIY cheaper
    versions will and are colormetric, DIY drop checkers are every
    bit as good as a colormetic cheapo test kit, with the KH
    reference solution they are better.

    So DIY and cheaper folks can take advantage of the method you
    suggested using the KH reference solution.

    The pH probe version is still a pH probe, just added a screw on
    DO membrane and have a company just adapt it for our specific
    use.

    It might cost 25-40$ more than a cheapo pH probe many folks use
    those also.

    You can DIY pH probes, thermo couplers(temp probes) and Redox
    probes etc as well, but most just buy those.

    Same deal here, you can adapt a pH probe, but most will not.
    It's nice to dIY, but it's also nice to have a company make it
    for those that do not want to deal with DIY.

    Basically it changes little in terms of the DIY/store bought
    issue we have **right now**, without the KH reference solution
    modification.

    " I'm
    guessing that this will be a nice lab instrument for the
    "serious"
    hobbyest, not the dilettante hobbyest like me. And, I don't
    have
    much doubt that it will work.

    Vaughn H."

    Naw....
    All we do is add a screw on cap that works on DO probes to a pH
    probe.

    We add different solution inside, baking soda vs KCL.
    This adaption should and dose work fine according to the
    companies that make the Teflon and poly membranes(I spoke to
    them as well as AM and Milwaukee).

    Regards,
    Tom Barr

    www.BarrReport.com
     
  2. VaughnH

    VaughnH Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 24, 2005
    Messages:
    2,891
    Likes Received:
    21
    Speaking of DIYing a pH probe: I have a Milwaukee portable pH meter that seems to still work ok, even after I neglected it for several months. The business end has a cap on it, where you are supposed to keep some water to keep that end wet at all times. I wonder if I could cut a hole in the end of the cap, cover the hole with semipermeable membrane, add the reference KH solution in the cap, reinstall the cap and use the assembly as a CO2 probe? See my crude sketch below: [​IMG]

    What is the major downside to cutting a hole in the cap that way? I could continue to store the probe with water in the cap, but it would let atmospheric CO2 enter the water.

    This post, this morning on APC, DIY Drop Checker - Page 13 - Aquatic Plant Central- aquascaping...a living art started me to thinking more about making a DIY CO2 probe. So, this afternoon I visited my local Home Depot, found their bin of Tyvek, helped myself to a 3" square of it (so, sue me, I swiped it!), and now I'm chomping at the bit looking for progress!
     
  3. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    I've reported you to the Sacramento police dept:p

    The rubber wet caps should work fine, the real issue is water leakage into the rubber cups at the membrane seal and also a rather obvious problem, the cost per unit of membrane.

    The AM and Milwaukee's are cheap and easy to just screw on.
    The plain membrane cost a lot more and you have to DIY it.

    Cost is important here as well.

    I want the item to marketed and sold, this will make it much more consistent and available and it should not much more than the cost of DO caps, about 8-10$

    Milwaukee Replacement Membrane for SM600 Dissolved Oxygen Mete

    And no KCL solution but a ref KH solution.

    Same type of deal, just made to fit a pH tip is all.

    A rubber DIY can be done, I have some new material coming in in a few days.

    I think epoxy on a rigid holder would work well for gluing it on. I used an O ring that was very tight. Then slip it down a bit and used a hypo needle to add the KH solution, you ALSO can do the filling before adding the O ring.
    This is a bit easier to do and less air bubbles inside.
    OSH, Home depot sells all sorts of O rings FYI.

    Still, 8$ ain't really saving you much for a DIY pH membrane ref probe adapter don't you think?

    Heck, that would not even pay for my time to go buy it and come back.

    Very simple upgrade and cheap, pre made for added consistency amongst it's users and availability., Folks can still DIY, mind you, but for saving $ on a DIY project when the item will not cost to start with?

    I have some sheets of PTFE membrane coming but have not figured a per unit price for each DIY.
    These are the most sensitive to changes in O2/CO2 properties.

    Artifical lungs are often studied and they use such membranes to exchange the CO2 and O2. Our lungs are still about 5x as good as the best artifical membrane.

    Regards,
    Tom Barr
     
  4. Bill

    Bill Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 23, 2005
    Messages:
    124
    Likes Received:
    0
    Are there any updates on the membrane material?

    What has been tried an what works best?

    Thanks,
    Bill
     
  5. VaughnH

    VaughnH Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 24, 2005
    Messages:
    2,891
    Likes Received:
    21
    We have been discussing using Kordon Breather bags as membranes on APC, and Dennis is trying the idea now. So far, without success. I expect to have a couple of the bags in a day or so, and I will experiment with them too.
     
  6. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    They are way too slow.

    Regards,
    Tom Barr
     
  7. VaughnH

    VaughnH Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 24, 2005
    Messages:
    2,891
    Likes Received:
    21
    That has been Dennis's experience so far, but he had a bubble of air above the KH water when he tried it, which would slow it down even more. The last he posted it was taking more than 3 hours for any color change to start up.
     
  8. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    I could have told you that to start with:)

    Use the poly membranes or the teflon.
    They are the best and the teflon tend to tougher and last longer, but are not as sensitive to changes.

    No need to suggest which I prefer:D

    I use a 10mm O ring, add the KH solution inside the space, the pH probe is 12 mm, so this makes a nice fit.

    I'm still not sure how long between ref solution replacements I need to do yet.
    It's still pretty new, they do make a ISE type CO2 probe that is similar, this is just a simpler cheaper DIY probe, the pre made CO2 probe runs about 395$.

    I prefer the 50$ ones and the 5 $ of materials.

    I sold off my old pH meter and replaced it with a nice Hanna pH/Redox/Temp/ISE meter for home use.
    So I have some interesting work ahead with respect to plant tanks.

    Regards,
    Tom Barr
     
  9. Bill

    Bill Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 23, 2005
    Messages:
    124
    Likes Received:
    0
    Hi Tom,

    I did a search and there are a lot of different poly membranes out there. Do you have any suggestions on a particular one and a source for them?

    Thanks,
    Bill
     
  10. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    The type of material used as a membrane also critically affects the performance of the system. Some have found that 0.001" to 0.0009" thick polyethylene, which is readily available from normal sandwich bags, works best for O2.

    Membranes that are more oxygen permeable, such as TFE Teflon or silicone, may be required for experiments at low temperatures or when a response rate faster than 4 seconds is desired. However, these materials will cause a greater motion artifact because the concentration gradient extends further into the bulk solution. TFE Teflon will also stretch and thin over time, causing the initial amplifier calibration to become incorrect. FEP Teflon usually works well, but is too stiff to wrap around the small electrode tip. Saran type wrap is not sufficiently permeable to oxygen to be useful as a membrane.

    The structure of sandwich bags tends to be poor even if the base material is the same.


    Regards,
    Tom Barr
     
  11. Patrice

    Patrice Guru Class Expert

    Joined:
    Jan 6, 2006
    Messages:
    159
    Likes Received:
    0

    Please tell me if I understain:
    I put kh ref solution in the cap
    make a hole in the cap
    put a membrane over the cap (sanwich bag?)

    and what?
    i do not understain "how this work".
     
  12. VaughnH

    VaughnH Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 24, 2005
    Messages:
    2,891
    Likes Received:
    21
    This works just like a "drop checker" works. The 4-5dKH water captured around the works of the pH probe absorbs CO2 from the tank water until it is in equilibrium with the tank water, meaning until it has the same ppm of CO2 as the tank water. That changes the pH of the 4-5dKH water according to the equation used to generate the table of ppm of CO2 vs pH and KH. So, the pH probe reading is a reading of the CO2 in the tank water. The cap for that probe would work as a compartment for the 4-5dKH water since it is already the right size and seals pretty well.

    I haven't tried this probe modification yet, and may not try it at all, since I have to cut a hole in the probe cap to do it. I don't yet have a gas permeable membrane to use, but an ordinary sandwitch bag isn't likely to work well, because even if it is gas permeable it would be too slow. A Kordon breather bag will work, but extremely slowly, which isn't usable for a hand held probe. You can also buy a package of membranes for an O2 probe from Cole-Parmer for a reasonable price and they might work fine. I just haven't gotten around to doing it yet. Tyvek is highly unlikely to work at all well, for the same reason - slow response.
     
  13. Patrice

    Patrice Guru Class Expert

    Joined:
    Jan 6, 2006
    Messages:
    159
    Likes Received:
    0
    great! i understain and want to try it. I have to assume it, I'me a gear freak, but I also have to assume that I am not ritch... yet :D

    so lets try that DIY

    so the first thing to get will be that membran. what do you think about Gore-Tex? would it work? I have an old jacket here so maybe I could use that.
    Other than this, that would be bether to have something more than just a membran hold with an elastic. What I have is a Hanna ph checker. do you think that milwakee cap showed previously on this post would fit on it? is the membran good on this?

    Also, i understain that the kh ref will absorbe the Co2 from the water. does it mean i will have to use new kh ref for each test?

    thx!
     
  14. Mr G

    Mr G Junior Poster

    Joined:
    Nov 24, 2006
    Messages:
    17
    Likes Received:
    0
    OK, Just a very quick post to say thanks to VaughnH and Tom for persuing the drop checker and PH probe methods .....
    I've recently been trying VaughnH's drop checker method Here and Tom suggested a more accurate way of mixing up the ref solution and am happy with the results.

    There are at least 4 other people over on that forum who are now giving the drop checker method a try. I'm posting off some ref solution this morning for one of them.
    I think many in the hobby will benefit from your efforts !

    I'm following the progress of the probe method and if I find anywhere in the UK where I can purchase some suitable membrane (without breaking the bank), I'll give it a try.

    Thanks again gents !
     
  15. VaughnH

    VaughnH Lifetime Charter Member
    Lifetime Member

    Joined:
    Jan 24, 2005
    Messages:
    2,891
    Likes Received:
    21
    YSI Meter Accessories And Replacement Membrane Kits - Cole-Parmer Catalog is a possibility for membranes, at a reasonable price. I just ordered a package of these last night, so I haven't tried them yet. The Cole-Parmer website is an international site, with "stores" for other countries, so these may be available in Europe or nearby also.

    To give credit where it is due: Dennis Dietz found this product and posted the link on APC.
     
  16. Patrice

    Patrice Guru Class Expert

    Joined:
    Jan 6, 2006
    Messages:
    159
    Likes Received:
    0
    thx for the link

    and about that kh ref solution, what do you use?
     
  17. Tom Barr

    Tom Barr Founder
    Staff Member Administrator Social Group Admin

    Joined:
    Jan 23, 2005
    Messages:
    18,458
    Likes Received:
    323
    Cooked/baked bakign soda: 400F for 45 min.
    Use about 2x what you think you need.

    10 grams may yield 6 grams after dehydrating.
    Sometimes 8 if you live in drier areas.

    I add 4.99 grams to 5 liters of DI water.
    This makes 40KH.

    Add 10mls of this to 90 mls of DI water to make a 4KH solution.

    Regards,
    Tom Barr
     
  18. Mr G

    Mr G Junior Poster

    Joined:
    Nov 24, 2006
    Messages:
    17
    Likes Received:
    0
    Just to say, I've tried this method of Tom's and it works very well and is easier than using a smaller volume of water and trying to adjust it with tiny amounts of bicarb. After diluting to 4dKH I double checked it with a KH test kit .... comes out at 4 !

    I can confirm Toms advice to use more baking soda than you think you'll need.
    I cooked off 7g and was left with less than 4.5g, so had to repeat the process.
    Bicarb is cheap .... so cook it up !

    Al
     
  19. Patrice

    Patrice Guru Class Expert

    Joined:
    Jan 6, 2006
    Messages:
    159
    Likes Received:
    0
    it's cooking! now I just need that membrane.
    thx lot for the advises
     
  20. dapellegrini

    dapellegrini Lifetime Charter Member
    Lifetime Member

    Joined:
    Oct 17, 2006
    Messages:
    111
    Likes Received:
    0
    Anyone ever figure out a good DIY cap/membrane for pH probes?

    I bought a drop checker and cooked up the kH solution for that. Works good...

    but onto the next project. I would like to build a cap for my pH Meter (Milwaukee Cheapo SMS122)... Wondering if anyone has found the "magic membrane". I would love to see someone spell this out clearly.
     